Synthetic Extracellular Volume Fraction Derived Using Virtual Unenhanced Attenuation of Blood on Contrast-Enhanced Cardiac Dual-Energy CT in Nonischemic Cardiomyopathy
작성자 관리자
작성일 21-10-13 11:35
조회수 87
Abstract
BACKGROUND
Current methods for calculating the myocardial extracellular volume (ECV) fraction require blood sampling to determine the serum hematocrit. Synthetic hematocrit and thus synthetic ECV may be derived using unenhanced attenuation of blood. By use of virtual unenhanced (VUE) attenuation of blood, contrast-enhanced dual-energy CT (DECT) may allow synthetic ECV calculations without unenhanced acquisition.
OBJECTIVE
The purpose of this study was to compare synthetic ECV calculated using synthetic hematocrit derived from VUE images and conventional ECV calculated using serum hematocrit, both of which were obtained by contrast-enhanced DECT, with ECV derived from MRI used as the reference standard.
METHODS
This retrospective study included 51 patients (26 men and 25 women; mean age, 59.9 ± 15.6 [SD] years) with nonischemic cardiomyopathy who, as part of an earlier prospective investigation, underwent equilibrium phase contrast-enhanced cardiac DECT and cardiac MRI and had serum hematocrit measured within 6 hours of both tests. A separate retrospective sample of 198 patients who underwent contrast-enhanced thoracic DECT performed on the same day for suspected pulmonary embolism and serum hematocrit measurement was identified to derive a synthetic hematocrit formula using VUE attenuation of blood by linear regression analysis. In the primary sample, two radiologists independently used DECT iodine maps to obtain the conventional ECV using serum hematocrit and the synthetic ECV using synthetic hematocrit based on the independently derived formula. The concordance correlation coefficient (CCC) was computed between conventional ECV and synthetic ECV from DECT. Conventional ECV and synthetic ECV from DECT were compared with the ECV derived from MRI in Bland-Altman analyses.
RESULTS
In the independent sample, the linear regression formula for synthetic hematocrit was as follows: synthetic hematocrit = 0.85 × (VUE attenuation of blood) – 5.40. In the primary sample, the conventional ECV and synthetic ECV from DECT showed excellent agreement (CCC, 0.95). Bland-Altman analysis showed a small bias of −0.44% (95% limits of agreement, −5.10% to 4.22%) between MRI-derived ECV and conventional ECV from DECT as well as a small bias of −0.78% (95% limits of agreement, −5.25% to 3.69%) between MRI-derived ECV and synthetic ECV from DECT.
CONCLUSION
Synthetic ECV and conventional ECV derived from DECT show excel lent agreement and a comparable association with ECV derived from cardiac MRI.
CLINICAL IMPACT. Synthetic hematocrit from VUE attenuation of blood may allow myocardial tissue characterization on DECT without the inconvenience of blood sampling.
Journal: American Journal of Roentgenology
Author
Na Young Kim, Dong Jin Im, Jong-Chan Youn, Yoo Jin Hong, Byoung Wook Choi, Seok-Min Kang, and Hye-Jeong Lee
DOI: doi.org/10.2214/AJR.21.26654
-
이전글
Evaluation of the Ostium in Anomalous Origin of the Right Coronary Artery with an Interarterial Course Using Dynamic Cardiac CT and Implications of Ostial Findings -
다음글
Comparison of Coronary Computed Tomography Angiography Image Quality With High-concentration and Low-concentration Contrast Agents: The Randomized CONCENTRATE Trial